|
[quote]原帖由 CPQinxu 于 2008-9-17 01:37 发表 ![]()
一般用用胜利足够了
专业的还是上台表吧
lala那里的台表很yy [/quo我们单位有纳伏表
那个更YY,明天贴张图
7位半的
Nanovoltmeter features
1
The Model 2182 is a 7 /2-digit high-performance digital nanovoltmeter. It has two input
channels to measure voltage and temperature. The measurement capabilities of the Model 2182
are explained in Section 2 of this manual (see “Measurement overview”).
Features of the Model 2182 Nanovoltmeter include:
* Ratio — Provides comparison readings between two voltage inputs. Ratio performs
V1/V2.
* Delta — Provides average difference of Channel 1 inputs. Delta performs
(V1t1–V1t2)/2.
* Enhanced Delta, Pulse Delta, and Differential Conductance — The following tests
can be performed when using a Model 2182/2182A with a Model 6220 or 6221 Current
Source:
— Delta - Uses a square wave output and a 3-point measurement algorithm to cancel
the effects of thermal EMFs.
— Pulse Delta (6221 and 2182A only) - Provides a pulse output and a 3-point (or
2-point) measurement algorithm for testing of temperature sensitive Device Under
Test (DUT).
— Differential Conductance - Uses a differential current output and a 3-point moving
average algorithm to perform differential measurements.
* mX+b and Percent — These calculations provide mathematical manipulation of
readings.
* Relative — Null offsets or establish baseline values.
* Buffer — Store up to 1024 readings in the internal buffer.
* Limits — Set high and low reading limits to test devices.
* Internal Scanning — Scan the two input channels of the Model 2182.
* External Scanning — Scan the channels or matrix points of Keithley Model 7001/7002
switching cards.
* Setup Storage — Two instrument setups (user and factory defaults) can be saved and
recalled.
* Analog Output — With analog output gain set to one, a full range input will result in a
1V analog output.
* Remote Interface — The Model 2182 can be controlled using the IEEE-488 interface
(GPIB) or the RS-232 interface.
* GPIB Programming Language — When using the GPIB, the instrument can be
programmed using the SCPI or Model 182 (DDCs) programming language.
* Closed-cover Calibration — The Model 2182 can be calibrated from either the front
panel or the GPIB.
[ 本帖最后由 veeco 于 2008-9-17 01:51 编辑 ] |
|